Howtek drum scanner DIY thread: maintenance, troubleshooting, mods
Old 09-20-2016   #1
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Howtek drum scanner DIY thread: maintenance, troubleshooting, mods

A place to discuss anything related to Howtek scanners.
  Reply With Quote

moved from Scanview thread...
Old 09-20-2016   #2
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
moved from Scanview thread...

Quote:
Originally Posted by onnect17 View Post
Yes. During auto calibration the Howteks will search for the density corresponding to the white point, before the A/D conversion. You should be able to see a couple of areas/strips 3/4" wide marked in the drum. The first one is typically covered by a white tape, used to calibrate the reflective work. I do not scan any reflective media so I removed it. The second one determine the white point for transparency mode. This area should be covered with the blank film. This helps a little if DPL (and log mode) is not available.
THANKS!

So, after WP calibration, the scanner internally works with min. density set to the min. density of the negative which leaves more bits to the actual data range of the negative film. Makes sense now. Will try it!

Quote:
Originally Posted by onnect17 View Post
Yes. The easiest way is to ask Aztek for a demo, so you can compare by yourself. After all, you are the one paying $1300.
I think I've asked for a demo but it wasn't available (at that time). Maybe they have a demo now (for DPL 8). Problem is, DPL8 is too expensive for me. I asked Aztek if DPL7 was still available and they made me an offer on DPL 7.86 Standard, but I'm still undecided (it's still $600).

Quote:
Originally Posted by onnect17 View Post
It's quite simple but you need the tool. I posted a drawing of it somewhere... can't remember now. I can post it again if you are interested.
The tool is probably not widely available (and expensive), right?

Below is a link to my "raw" scan from Silverfast later, maybe you can guess from it if the results could be improved with calibration of the light path or it's just what can be expected from scanning a negative film in Silverfast (I scan 35mm negs with 4000dpi and 19 aperture). Do you have any scans available made with Silverfast?


Link to full res raw scan.

This is what I get after inverting (100% crop). If the negative is dense (like those overexposed Fuji400H), things get even worse...



(unfortunately, I don't have any sharp negative raw files at hand - this is from a plastic Instamatic lens)
__________________

  Reply With Quote

Old 09-20-2016   #3
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Quote:
Originally Posted by brbo View Post
THANKS!
So, after WP calibration, the scanner internally works with min. density set to the min. density of the negative which leaves more bits to the actual data range of the negative film. Makes sense now. Will try it!
Ideally, you could pull-develop a roll of Ektar 100 by 1/2 stop or just cut time by 20 or 30 secs. The goal is to use a strip with min densities around 0.1 lower than the typical min densities for the C41.

Quote:
Originally Posted by brbo View Post
The tool is probably not widely available (and expensive), right?
It's not, actually. I will post details later. All it does is to hold a piece of 1/4" acrylic (similar to the drum thickness) in front of the focusing lens.

Quote:
Originally Posted by brbo View Post
I'll post a link to my "raw" scan from Silverfast later, maybe you can guess from it if the results could be improved with calibration of the light path or it's just what can be expected from scanning a negative film in Silverfast (I scan 35mm negs with 4000dpi and 19 aperture). Do you have any scans available made with Silverfast?
I have to dig for a old sample later. Not sure if I can find anything in ISO 400.
In any case you should verify the alignment of the light source, regardless of the app you using to scan, especially important with focused illuminators. Also, 19 microns is too open IMHO. It's like 1000 dpi optical.
  Reply With Quote

Howtek 4000/4500 alignment tool diagram
Old 09-20-2016   #4
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Howtek 4000/4500 alignment tool diagram

Here's a drawing of the tool. Feel free to ask any questions.

https://1drv.ms/i/s!AugK5NQO79gPgn7NIYn5kvc12XDr
  Reply With Quote

Old 09-20-2016   #5
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
Oh, the questions will follow...

So this tool attaches to something(1) and you do the alignment of something(2) and then you observe the success of the alignment through something(3)?

Now I need a manual and start reading... Does this tool do anything else than act as a substitute for the drum, but with this tool the drum is not obstructing access to the point where you do the adjustments?

Ok, I really should start looking for and then reading the manual...
__________________

  Reply With Quote

Old 09-20-2016   #6
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Quote:
Originally Posted by brbo View Post
Oh, the questions will follow...

So this tool attaches to something(1) and you do the alignment of something(2) and then you observe the success of the alignment through something(3)?

Now I need a manual and start reading... Does this tool do anything else than act as a substitute for the drum, but with this tool the drum is not obstructing access to the point where you do the adjustments?

Ok, I really should start looking for and then reading the manual...
Sorry, I assumed you were familiar with the Service Manual. The alignment procedure is described in pages 6-1 to 6-3. Check the group in yahoo dealing with this scanners to download a copy. I think is named Scan-HiEnd or something like that.

The alignment is done without a drum mounted so you can access the illuminator.
  Reply With Quote

Old 09-20-2016   #7
meloV8
Registered User
 
meloV8 is offline
Join Date: Sep 2014
Posts: 54
brbo, quick raw invertion in silverfast hdr :

http://www.filedropper.com/h4500silverfastcolor

To me looks ok
  Reply With Quote

Old 09-20-2016   #8
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
Yes, Silverfast's Negafix... I don't know if I want to go there. There are times that I almost like the results, but I always come back to ColorPerfect. I can make the final results look like Silverfast (regarding noise) and still have some details in the highlights...




But still, as onnect17 pointed out, this is at 19microns. I'll try to scan some Ektar at 6 and 13microns. Last time I did that bright sky looked like my childrens' room after a week long Lego session
__________________

  Reply With Quote

Old 09-21-2016   #9
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
After reading the manual about the optical alignment I don't think I need the tool to check the alignment. I'd only need it to make adjustments.

Lets hope the check reveals nicely centered light and that all that it takes is paying the Aztek price

BTW, is there any point in changing the lamps even if I don't notice any streaks? I have no idea how old they are...
__________________

  Reply With Quote

Old 09-21-2016   #10
quejai
Registered User
 
quejai is offline
Join Date: Jul 2012
Posts: 103
Quote:
Originally Posted by brbo View Post
After reading the manual about the optical alignment I don't think I need the tool to check the alignment. I'd only need it to make adjustments.

Lets hope the check reveals nicely centered light and that all that it takes is paying the Aztek price

BTW, is there any point in changing the lamps even if I don't notice any streaks? I have no idea how old they are...
I think the idea behind lamp changing is that the colour temperature / emission spectrum of the lamp could change very slightly with time, but that change might be considerable compared to the very subtle colour graduations in the film, especially negative film.
I think you should be able to get around it by using a colour calibration slide now and then, then you should be fine until it burns out.

There might be other reasons too, but I'm not sure of them.
  Reply With Quote

Old 09-21-2016   #11
meloV8
Registered User
 
meloV8 is offline
Join Date: Sep 2014
Posts: 54
I tried the ColorPerfect, but I could not obtain good results. I use Silverfast to reverse negative, then Lightroom for more precise work.
  Reply With Quote

Old 09-21-2016   #12
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Quote:
Originally Posted by brbo View Post
After reading the manual about the optical alignment I don't think I need the tool to check the alignment. I'd only need it to make adjustments.

Lets hope the check reveals nicely centered light and that all that it takes is paying the Aztek price

BTW, is there any point in changing the lamps even if I don't notice any streaks? I have no idea how old they are...
Remember, the purpose of the tool is not only check if the illuminator projects a centered beam. The width of the beam is also important and the acrylic in the body of the drum affect the optical path.
  Reply With Quote

Old 09-22-2016   #13
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
Quote:
Originally Posted by onnect17 View Post
Remember, the purpose of the tool is not only check if the illuminator projects a centered beam. The width of the beam is also important and the acrylic in the body of the drum affect the optical path.
Do you have any control of the width of the beam? Manual only discusses the centering of the beam.

Read some vague comments (like in this thread) that you can get by without alignment tool. Maybe I can still move the illumination lens with the drum mounted (I have one drum with heavy crazing so I'm not afraid of scratching it)? Which two screws need to be unscrewed to adjust the beam?



I also found that I'm missing the screw that's supposed to hold the FORI assembly in place.



The thing doesn't move and by the looks of it sits tight in position, but I should probably find a screw to secure the FORI. Anyone have a picture of the screw?

I played a bit with scanning with negative placed over the calibration strip. Maybe a bit better but still pretty bad. Ektar raw scan. And after inverting (with heavy noise reduction).
__________________

  Reply With Quote

Old 09-22-2016   #14
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Quote:
Originally Posted by brbo View Post
Do you have any control of the width of the beam? Manual only discusses the centering of the beam.
The width of the beam is controlled by screwing in/out the illuminator lens.

Quote:
Originally Posted by brbo View Post
Read some vague comments (like in this thread) that you can get by without alignment tool. Maybe I can still move the illumination lens with the drum mounted (I have one drum with heavy crazing so I'm not afraid of scratching it)? Which two screws need to be unscrewed to adjust the beam?
That's the tool I got. Check the yahoo forum for some pics. (hard to post them here due to size restrictions)

I still recommend you to do it without a drum. Place the scanner OFFLINE. Then select FOCUS and the option to MOVE CARRIAGE.

Half way in the arm you can find one or two allen screws that allows you to move in/out and rotate the whole illuminator head.

Quote:
Originally Posted by brbo View Post
I also found that I'm missing the screw that's supposed to hold the FORI assembly in place. The thing doesn't move and by the looks of it sits tight in position, but I should probably find a screw to secure the FORI. Anyone have a picture of the screw?
Some scanners came with a knurled head screw and others with an allen screw. Most likely you have the second option. Hard to see from outside.

Quote:
Originally Posted by brbo View Post
I played a bit with scanning with negative placed over the calibration strip. Maybe a bit better but still pretty bad. Ektar raw scan. And after inverting.
Wait until you finish the alignment and test it again.
  Reply With Quote

Old 09-22-2016   #15
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
Thanks! I got the drawing of the tool that you posted here and in ScanHi-End group. I'd need to investigate where I could have one made... For now, I'll have to try with the check-adjust-check-adjust method first.
__________________

  Reply With Quote

Old 09-23-2016   #16
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
What is the optimal shape and position of the light beam (and why)?

Centered, because the center is the brightest and has least aberrations?

As narrow as possible (but still covering the largest aperture that will be used)?


If I could get the alignment tool and since you have to remove the FORI to mount it, is there any point in remounting the FORI after calibration? I don't plan to scan any reflective material. In theory, it adds (unwanted?) reflections, right? On the other hand it would leave analyser lens unshielded and somewhat exposed to mechanical damage.
__________________

  Reply With Quote

Old 09-23-2016   #17
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Quote:
Originally Posted by brbo View Post
What is the optimal shape and position of the light beam (and why)?
Centered, because the center is the brightest and has least aberrations?
As narrow as possible (but still covering the largest aperture that will be used)?
Correct, but the procedure described in section 6.1 of the service manual is only intended to maximize the amount of light reaching the aperture wheel. It does not address the alignment of the analyzer lens to find the optical center. To do so you could use a different tool that uses a small cross-point laser, like the one in this link--> http://www.largeformatphotography.in...=1#post1240695

The shape of the beam, as displayed in the view port, should look like the one shown in Section 6.5, page 6-7.

Quote:
Originally Posted by brbo View Post
If I could get the alignment tool and since you have to remove the FORI to mount it, is there any point in remounting the FORI after calibration? I don't plan to scan any reflective material. In theory, it adds (unwanted?) reflections, right? On the other hand it would leave analyser lens unshielded and somewhat exposed to mechanical damage.
The FORI also acts as an aperture. So without it too much unnecessary light would get in. I taped (using black matte tape) a square aperture of approx. 1/8" in front of the FORI. Happy with the results.
  Reply With Quote

Old 09-23-2016   #18
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
This was before adjustment:




After:



Not perfectly centered, but better than before. First test scan didn't show any massive improvement.


The light in viewport after adjusting looks like this:




Not quite centered and not really the same as manual suggests (the cener patch is too big and I don't see the second dark ring):




Hmm, getting the shape right will probably be harder than centering the light on the aperture wheel without the alignment tool...
__________________

  Reply With Quote

Old 09-24-2016   #19
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
That's was quite off at the aperture wheel. Enough to see less color aberrations in the image, assuming the analyzer lens is in proper alignment.

To reduce the width of the beam try screwing in/out the illuminator lens. Also check the spot projected by the lamp on the entrance to the fiber. Make sure is the smallest possible but still covering the 5mm fiber bundle. It may require loosing the allen screws supporting the frame of the lamp. A ND filter could help protecting your eyes here.
  Reply With Quote

Old 09-24-2016   #20
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
Quote:
Originally Posted by onnect17 View Post
That's was quite off at the aperture wheel. Enough to see less color aberrations in the image, assuming the analyzer lens is in proper alignment.

To reduce the width of the beam try screwing in/out the illuminator lens. Also check the spot projected by the lamp on the entrance to the fiber. Make sure is the smallest possible but still covering the 5mm fiber bundle. It may require loosing the allen screws supporting the frame of the lamp. A ND filter could help protecting your eyes here.
Is there any way to adjust the analyser lens? Can't see it mentioned in the manual.

If I get good centering on the aperture wheel but off-centered image in viewport this could also be because of flip-up mirror? This is non-issue, but if the cause is in the analyser lens misalignment...

Could you explain how to check if the analyser lens is positioned correctly? Unfortunately, I don't understand how one would use the cross-point laser for that. Thanks!
__________________

  Reply With Quote

Old 09-24-2016   #21
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Quote:
Originally Posted by brbo View Post
Is there any way to adjust the analyser lens? Can't see it mentioned in the manual.

If I get good centering on the aperture wheel but off-centered image in viewport this could also be because of flip-up mirror? This is non-issue, but if the cause is in the analyser lens misalignment...

Could you explain how to check if the analyser lens is positioned correctly? Unfortunately, I don't understand how one would use the cross-point laser for that. Thanks!
The analyzer lens is quite similar to a microscope objective, even share the same old thread. You can remove it easily with the help of a rubber band. Imagine if you make a similar piece and instead of a couple of achromats elements you insert a small laser pointing inwards. That will show the sweet spot of the analyzer lens.

The analyzer lens is mounted on a linear stage that can be adjusted.
  Reply With Quote

Old 09-25-2016   #22
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
The laser thing will have to wait as I don't have it and don't really understand what I'd have to do...

Anyway, I swapped reflective and transmissive lamps and cleaned analyser lens. No noticeable improvement. Either there is a problem somewhere else or this is "par for the course". Maybe Howtek 4500 just can't be compared to a good CCD scanner (I also have Minolta 5400) when scanning negatives. I'm a bit hesitant to touch the mirror and filters inside the optical box to clean the dust off. Is this something that can safely be done with cotton swabs and lens cleaning solution?

For "grainy/discolored image/colored streaks" further suggestions from manual include replacing PMTs, VAA, VDS, pre amplifier board... So I'd probably be out of luck if that is the source of my problems. My scans of slides look good to me, though. If I push shadows in PP there's some noise, but that's probably normal. Can field service test program give more answers? Can the FST4500.exe be run directly from within WinXP?

I managed to put together an old PC and get Trident 3 going in demo mode. Scans of negatives have considerably less color noise (still look a fair bit "harsher" than CCD scans, but this was all dry mounted on a crazed drum). I guess Trident is using internal 12bits to only get the actual range of the negative. I guess DPL Pro would be roughly the same then so I have a kind of DPL demo now
__________________

  Reply With Quote

Old 09-26-2016   #23
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Quote:
Originally Posted by brbo View Post
Maybe Howtek 4500 just can't be compared to a good CCD scanner (I also have Minolta 5400) when scanning negatives. I'm a bit hesitant to touch the mirror and filters inside the optical box to clean the dust off. Is this something that can safely be done with cotton swabs and lens cleaning solution?
It's really hard to compare scanners especially if are using different technologies. A 4500 in good shape produces an excellent scan, with clean color channels than no CCD based can get even close. I owned a few of them, including a Nikon V ED and an Imacon 343. Sold them. Never got used to the idea of having lower resolution in the borders, or uneven focus, flare, etc.

It's fine to clean the optics but be careful with the aperture wheel. You can inspect the holes using a microscope.

Quote:
Originally Posted by brbo View Post
For "grainy/discolored image/colored streaks" further suggestions from manual include replacing PMTs, VAA, VDS, pre amplifier board... So I'd probably be out of luck if that is the source of my problems. My scans of slides look good to me, though. If I push shadows in PP there's some noise, but that's probably normal. Can field service test program give more answers? Can the FST4500.exe be run directly from within WinXP?
IMHO the "achilles heel" of the Howteks is the underpower light source. The alignment procedure would get the best of the original hardware but you can also improve the output installing a more efficient lamp, like RGB LED or at least placing a filter UV-IR at the entrance of the fiber. Also replacing the PMT with other models with better performance, like the R3896, R928, etc. With this changes I use the 6 microns aperture all the time, without issues.

I run the FST4500 from a floppy with DOS or directly from a DOS HD partition, using a PC or Notebook. Not sure if I can run it from XP. The section to test the PMTs could give you an idea of their condition.

Quote:
Originally Posted by brbo View Post
I managed to put together an old PC and get Trident 3 going in demo mode. Scans of negatives have considerably less color noise (still look a fair bit "harsher" than CCD scans, but this was all dry mounted on a crazed drum). I guess Trident is using internal 12bits to only get the actual range of the negative. I guess DPL Pro would be roughly the same then so I have a kind of DPL demo now
I think Trident uses the Log mode by default, so the scans of C41 should look a little better.
  Reply With Quote

Old 09-28-2016   #24
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
Yes, flare in CCD scanners ruins it for me. Film flatness, too. I had a Microtek 120tf scanner that was AMAZING (for a CCD scanner) in regard to those two issues, though. The flare was more than what I get with my Howtek, but very managable. Any Nikon Coolscan would look like a joke in a scenario when you scan with sprocket holes included...






My Minolta 5400 with multisampling can match DMax of my Howtek (but not DRange). My Epson 4990 can match the size of largest film that can be scanned.... I'm sure a drum scanner in good condition can outdo them all in every characteristic, so I'm trying to get to the bottom of my problem with scanning negatives.


I run FST4500.exe, everything seem to be "OK". But this tests would probably need a person that is able to interpret the numbers to get past the "OK/Fail" level of troubleshooting in order to get to the part of the scanner that needs to be checked/improved. So here are screens of tests that I thought might be of relevance with the issue:









The test was done before I changed the lamp for a new one (I don't see any change in scans) and before I installed an uv/ir cut filter that I had at hand (I can see a small change):




I'll run the test again and see how uv/ir filter affects the numbers.

BTW, I'm curious about the reason behind the lamp shining into the optic fibre at an angle.


Oh, one more thing. What is the purpose of the battery in this scanner? Mine is dead as the scanner boots with "Checksum error" and then loads defaults. Could/would any of the pre-scanning routines be skipped if the battery was working?
__________________

  Reply With Quote

Old 09-28-2016   #25
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
The test I use the most is the PMTs. It should be close to the max, 4095. The rest is just a guess. Some basic info if you press F1 for help. Here are some pics of my tests here for comparison:

https://1drv.ms/f/s!AugK5NQO79gPgxBihXVwLijEBICW

You should replace the motherboard battery in any case because some users reported stability issues with it.

If I understand correctly, the boot code is in EPROM on the motherboard. In some point loads an updated version from flash memory located in one of the cards. I made copy of all the firmware, just in case.

I am curious about the CHECKSUM error. It would be helpful if you take a video showing it.
  Reply With Quote

Old 09-28-2016   #26
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
Thanks for screenshots. My numbers are not that far off. Green channel is lagging in some tests, but doubtful that it should make much difference.

I'll make a video of boot process when I get home. From memor it goes like "Starting scanner" -> "Checksum error. Loading defaults" -> "Soft reboot" -> "Ready to scan. Firmware R813".
__________________

  Reply With Quote

Old 09-28-2016   #27
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
Here is the link to the video of the scanner boot sequence.

I also ran the FST with UV/IR cut filter installed. The PMTs test shows lower values in green channel. In repeated tests values range from 3900 to 4090.

Found upt that the closer I move the filter to the fiber the higher the value. I've now removed the filter.
__________________

  Reply With Quote

Old 09-28-2016   #28
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Quote:
Originally Posted by brbo View Post
Here is the link to the video of the scanner boot sequence.

I also ran the FST with UV/IR cut filter installed. The PMTs test shows lower values in green channel. In repeated tests values range from 3900 to 4090.

Found upt that the closer I move the filter to the fiber the higher the value. I've now removed the filter.
The Flash version seems to be loading fine. Perhaps the battery is dragging down the 5v in the MB during the initial seconds, which is not a good thing anyways. Where are you located? I remember buying a pack of 5 batteries a few years back for not much money. Also make sure the scanner is connected to the same outlet used by the PC.

The UV/IR filter should be hold against the entrance of the fiber. Check for ebay item number 281437567116.
  Reply With Quote

Old 09-28-2016   #29
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
I'm from Slovenia (EU).

The scanner and the PC are on the same outlet and have the same surge protector between. Although the manual recomends an UPS. Wonder if that an actual requirement or just over-the-top suggestion.

If the scanner is shut down and powered back on within a minute or so there is no checksum error.
__________________

  Reply With Quote

Old 09-28-2016   #30
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Quote:
Originally Posted by brbo View Post
I'm from Slovenia (EU).

The scanner and the PC are on the same outlet and have the same surge protector between. Although the manual recomends an UPS. Wonder if that an actual requirement or just over-the-top suggestion.

If the scanner is shut down and powered back on within a minute or so there is no checksum error.
I am sure you are familiar with how clean the power is in your area. If you are located in a residential area then should be fine. However, if you have a few business around using heavy machinery then protection is recommended. UPS is not a must.

If you do not get an error within a minute then it seems like a MB battery issue. Easy to replace. eBay item 351792398716 should do the job.
  Reply With Quote

Old 09-30-2016   #31
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
Thanks! Ordered the battery. I hope there won't be soldering involved when replacing...

Anyway, I've done/checked what I was able by myself at this moment. I'm probably one aperture away from what I would like to have at scanning negative film. I'd be totally happy even with 13m for fine grained neg film, but with the most negs I've tried so far I needed 19m to keep noise in check.

So... I broke down and ordered DPL7. I hope it will help with negative film. Slide film is slowly (but much faster than I would want) going away and Kodak Vision3 stuff is great and not so expensive (even dirt cheap if you buy short-ends).
__________________

  Reply With Quote

Old 09-30-2016   #32
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
Quote:
Originally Posted by brbo View Post
Thanks! Ordered the battery. I hope there won't be soldering involved when replacing...
Unfortunatelly yes. And be prepare to spend a lot of time removing all the cards and connectors to reach the motherboard. Take pictures before you start!

Quote:
Originally Posted by brbo View Post
Anyway, I've done/checked what I was able by myself at this moment. I'm probably one aperture away from what I would like to have at scanning negative film. I'd be totally happy even with 13m for fine grained neg film, but with the most negs I've tried so far I needed 19m to keep noise in check.
"one aperture away" and "one aperture at a time" :-)

Quote:
Originally Posted by brbo View Post
So... I broke down and ordered DPL7. I hope it will help with negative film. Slide film is slowly (but much faster than I would want) going away and Kodak Vision3 stuff is great and not so expensive (even dirt cheap if you buy short-ends).
It should help but don't expect miracles. IMHO, it takes some of the hardware changes previously mentioned to really get down to 6 microns.
Good luck!
  Reply With Quote

Old 11-01-2016   #33
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
A couple of links to videos showing a basic test of the balance in the 7500. If operating properly, the liquid inside a small container on top of the scanner should not show any sign of vibrations, at both, slow and fast speed.

https://1drv.ms/v/s!AugK5NQO79gPgxdyS6e_KmOkGe7j

https://1drv.ms/v/s!AugK5NQO79gPgxgd-RqNCXZdBpuP
  Reply With Quote

Old 01-05-2017   #34
Fixcinater
Never enough smoky peat
 
Fixcinater is offline
Join Date: Dec 2013
Location: Portales, NM, USA
Posts: 377
Over the past months, I've been working on getting a D4000 setup running. It came with two drums, I've been using the worse condition one (few larger scratches on it, haven't tried polishing it at all) to get things started. The other nicer condition drum is safely packed away until I can put it to better use.

Current setup:

Mirror Drive Doors Mac G4 tower with OS 9.2.2 and 10.4.11
Adaptec 2906 SCSI PCI card
SCSI cable that came with the scanner
D4000 scanner on R535 firmware

Luckily, the MDD tower is one that allows booting directly into OS9, the last version with built in firewire 800 ports does not allow it.

So far, I've swapped bulbs from reflective to trans side which helped with some dimming during scanning leading to lines. The lead screw looks dry, should I lube it or not? I've seen both ways of thinking.

Issue 1: I have just updated the firmware on the Howtek to R535 (was 531), and now I cannot move the carriage through the Howtek's panel. It seems that this should not have been lost in the firmware update, but can anyone confirm that R535 does not allow manual carriage movement and AF and aperture choice? I was able to update the FW through the small app provided on Aztek's website along with the FW file itself. I only have tried this once, would it be worth trying again?

Issue 2: Silverfast 6.6.2r5 (demo version until I decide if I should invest in DPL or Silverfast) will talk to the scanner and make scans in OS 10.4.11. None of the demo software downloaded from Aztek's website work in OSX through Classic, which is disappointing but somewhat as expected. When booting directly into OS 9, Aurora will see the scanner but will not scan anything. Trident v3.5.5 and v4.0 will not see it at all, and Polaris will scan but none of the produced files are viewable, they turn into .exe files when moved to OS X or another Mac computer. Should I just quit wasting time trying to get those demo programs to work and buy Silverfast? The scanner came with an early version of Aurora, but no dongle so I can't access full version anyway.

Issue 3: The scanner came with two Howtek labeled drum mounting stations, but both of the waxy rubber pressure spools have literally melted off the spindles. They were kept in a garage in southern California and the door was left open, leaving a pool of waxy rubber on the floor. I've seen some pipe insulation wraps used on DIY mounting stations, are there any other suggestions?

Issue 4: Seems to be a large amount of vibration compared to the D7500 video posted just above. Should the D4000 be held to the same standard of smoothness? Maybe related: Scanner had a cracked old belt, I replaced with one that came with it (soft and supple, never installed before) as well as a new o-ring on the mounting spindle. How do I know what proper belt tension is? I can see the allen heads for adjusting the motor fore/aft on the carriage to set tension. Seems a bit loose now but I don't want to burn any bearings up.

Lastly, I just want to say thank you to all those that have posted their problems and solutions and those that have helped diagnose things, I never would have been able to get it going as far as I have without the existing discussions being able to be referenced.
  Reply With Quote

Old 01-06-2017   #35
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
The lead screw in the 4000 has a solid coating of teflon (PTFE). It's save to use a lubricant with a perfluoropolyether (PFPE) base and PTFE as a thickener. Check some thread in the large format forum about it. I use it in all the howteks because all of them use a plastic anti-backlash nut system.

Issue 1 - I have to go to the basement and check, but if I remember correctly, just follow the procedure to focus manually. That should allow you to move the carriage left and right. Be aware, there is a time out that returns the carriage to home after a while.

Issue 2 - Regarding Trident, I think you also need the scanner driver (version 2.2)

Issue 3 - Check the Hi-End scanner yahoo group for solutions related to the rollers in the mounting stations.

Issue 4 - The driving belt should not be issue here with the vibration. The o-ring in the mounting spindle and the pads at the other end of the drum could be the source. I use 6 or 8 pads instead of 3. Also the bearings could be the source.

Good luck and let us know how it goes.
  Reply With Quote

Old 01-08-2017   #36
Fixcinater
Never enough smoky peat
 
Fixcinater is offline
Join Date: Dec 2013
Location: Portales, NM, USA
Posts: 377
Thanks for the quick reply!

I rebooted scanner by pressing all three buttons and the firmware options became available again. Not sure what was happening there as I couldn't access any of those menus but after the reset I can move the carriage and change aperture, etc with the new firmware. So, I cleaned up underneath the carriage home position and called it a night.

Will update on the others as I make progress. Again, thank you for taking the time to help out!
  Reply With Quote

Old 01-08-2017   #37
onnect17
Registered User
 
onnect17 is offline
Join Date: Dec 2009
Posts: 74
You are welcome!
  Reply With Quote

Old 01-17-2017   #38
Fixcinater
Never enough smoky peat
 
Fixcinater is offline
Join Date: Dec 2013
Location: Portales, NM, USA
Posts: 377
Minor update after trying a few things to reduce the lines I have been seeing some lines along drum rotation axis. I noted that I only saw it when set to 2000DPI or higher.

-Lubed the lead screw (was totally dry), scanning is quieter - no change.
-Changed the bulb out for yet another one that came with the scanner that was labeled "still usable" - no change.
-Confirmed bulb is shining on end of fiber pack - no adjustment necessary, no change
-Couldn't tell where the drum pads were supposed to be located, checked the other drum that came with the scanner (has been in the factory pouch in safe-keeping), saw where the three pads are and felt rather stupid.

Next step is confirming that having ANY pads helps, which I can only assume it will...and ordering more pads so I can go 6 or 9 pads on each drum.
  Reply With Quote

Old 01-17-2017   #39
brbo
Registered User
 
brbo's Avatar
 
brbo is offline
Join Date: May 2011
Posts: 987
Describe or show us the lines you are getting in the scans. Is the calibration part of the drum perfectly clear/clean?
__________________

  Reply With Quote

Old 01-18-2017   #40
Fixcinater
Never enough smoky peat
 
Fixcinater is offline
Join Date: Dec 2013
Location: Portales, NM, USA
Posts: 377
Attached are two files from small scans I just did. Ambient room temp is fairly cold, scanner has been turned on for ~2hrs at this point.

All settings between files are same, I resized with CS3 to same pixel width.

EDIT: This is with a transformer style power supply for the scanner. I can still clearly see variation in the lamp brightness, sometimes seems there is a pattern to the variance and sometimes it just dims quickly and then back to norm.
Attached Images
File Type: jpg [email protected] (47.5 KB, 9 views)
File Type: jpg [email protected]_DPI.jpg (51.2 KB, 9 views)
  Reply With Quote
Reply


Thread Tools Search this Thread
Search this Thread:

Advanced Search
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

vB code is On
Smilies are On
[IMG] code is On
HTML code is Off



All times are GMT -8. The time now is 06:53.


vBulletin skin developed by: eXtremepixels
Powered by vBulletin® Version 3.6.8
Copyright ©2000 - 2017, Jelsoft Enterprises Ltd.

All content on this site is Copyright Protected and owned by its respective owner. You may link to content on this site but you may not reproduce any of it in whole or part without written consent from its owner.